Q&A Forum

What is process measurement assurance and do small labs need it?

Asked by Metrology Atlas Editorial · 2 weeks ago · ✓ has an expert-verified answer

NIST SOP 30 describes 'process measurement assurance' with check standards and control charts. We're a two-person lab — is this overkill between our annual calibrations?

1 Answer

Metrology Atlas Editorial ✓ Verified expert answer 2 weeks ago

It scales down well and is arguably MORE valuable for a small lab, because an annual external calibration alone tells you nothing about the other 364 days.

Minimum viable program per SOP 30 / GLP 1: pick one stable check standard per measurement area, measure it with your normal process weekly or per batch, plot on a control chart with limits from your accumulated standard deviation, act on out-of-control points (investigate before releasing results). Time cost: minutes per week.

What it buys you: drift detection between calibrations, evidence for auditors that results stayed in control since the last calibration (limits the recall scope if something fails as-found), and a live Type A estimate feeding your uncertainty budget. ISO/IEC 17025 (7.7.1) effectively expects it — 'monitoring the validity of results'.

Know the answer?

Sign in to post an answer under your own name.

SIGN IN TO ANSWER

Have a similar measurement problem?

Ask the community — anonymously, with as much technical detail as you can.